"Patient and Surgeon Radiation Exposure: Comparison of Standard and Mini-C-Arm Fluoroscopy"
by Brian D. Giordano, MD, et al.

Commentary & Perspective by
Matthew J. Allen, Vet MB, PhD*,
Ohio State University, Columbus, Ohio

Posted February 2009

Despite the widely recognized dangers of ionizing radiation, orthopaedists routinely place themselves in harm's way during the course of their daily professional activities through the use of fluoroscopy. A number of studies have attempted to quantify the risk of radiation exposure from the perspectives of both surgeon and patient. The consensus from these studies has been that the doses absorbed by the patient and the physician are relatively low as compared with current recommended annual exposure limits1 and may be further reduced through safe working practices that are based on the ALARA (as-low-as-reasonably-achievable) principle2. Mini-C-arms also offer an advantage since they permit excellent image acquisition with concomitant reductions in the radiation dose as compared with standard C-arms3.

The authors of the current study have recognized two important limitations in the application of these findings to current orthopaedic practice. Firstly, most of the available literature on radiation exposure from C-arms or mini-C-arms is based on imaging of small joints, such as the hand or wrist. In a practical setting, there is a growing trend toward the use of fluoroscopy to image larger bones and joints, for example in spine surgery and minimally invasive fracture repair. Since the size of the specimen that is to be imaged has a direct effect on the radiation dose required for imaging, the authors sought to quantify the total radiation doses to which patients and physicians are exposed during large-joint imaging. The second issue that they raise is that of inadequate training in either radiation safety or in the operation of the C-arm. Suboptimal positioning of the target tissue within the C-arm has the potential to increase both the direct radiation dose to the patient and the scattered radiation dose to the surgeon and assistants.

In an attempt to address these concerns, the authors designed a simple experiment in which they quantified radiation exposure with use of either a full-size C-arm or a mini-C-arm during fluoroscopic examination of a cadaveric human ankle joint. Dosimeters were placed on the skin at the level of the ankle joint (to measure the surface dose received by the patient) and around the x-ray source and image intensifier in order to quantify x-ray scatter to which the surgeon might be exposed.

The results from this study confirm the importance of correctly positioning the limb within the C-arm. For a full-sized C-arm with the ankle joint positioned 2 inches from the x-ray tube (considered the worst-case scenario), the "patient" was exposed to a skin dose of 8988 mrem. Scattered radiation doses varied from <10 mrem to 38 mrem, depending on the precise location relative to the C-arm. If the same specimen configuration was used with the mini-C-arm, the patient dose was 3912 mrem, a decrease of 56% relative to the full-size unit, while the scattered dose ranged from <10 mrem to 18 mrem.

With the ankle joint positioned immediately above the image intensifier (considered the best-case scenario), the patient dose with a full-size C-arm was 850 mrem and the scattered dose ranged from <10 mrem to 17 mrem. Use of the mini-C-arm reduced the patient dose to 305 mrem but had no significant effect on the scattered dose.

The results from this study demonstrate the range of measurable radiation exposure to both the patient and the surgeon, depending on the type of C-arm used and the technique. In the scenario that was modeled, the exposure rate was on the order of 38 mrem for a five-minute exposure. These data are consistent with exposure rates measured directly on the hands of surgeons who are performing hand surgery under mini-C-arm fluoroscopy (5 to 80 mrem per case)4. Positioning of the imaged specimen closer to the radiation source increased patient exposure but had relatively less effect on surgeon exposure. For any given specimen orientation, radiation doses were reduced with a mini-C-arm as compared with a standard C-arm.

The scattered radiation doses measured in this study were relatively low and were roughly equivalent to the surface dose from a routine chest x-ray (approximately 20 mrem). However, repetitive exposure of surgeons to even low levels of scattered radiation remains a cause for concern. Medical personnel involved in fluoroscopic procedures should be trained and certified both in the principles of radiation safety and in the approved operating procedures for the specific C-arm that they are using. The authors are to be commended for this detailed and thoughtful study that reminds all of us of our obligations with regard to the safety of the patient and the attending medical team.

*The author did not receive any outside funding or grants in support of his research for or preparation of this work. Neither he nor a member of his immediate family received payments or other benefits or a commitment or agreement to provide such benefits from a commercial entity.

References

1. Singer G. Occupational radiation exposure to the surgeon. J Am Acad Orthop Surg. 2005;13:69-76.
2. Brateman L. Radiation safety considerations for diagnostic radiology personnel. Radiographics. 1999;19:1037-55.
3. Athwal GS, Bueno RA Jr, Wolfe SW. Radiation exposure in hand surgery: mini versus standard C-arm. J Hand Surg [Am]. 2005;30:1310-6.
4. Singer G. Radiation exposure to the hands from mini-C-arm fluoroscopy. J Hand Surg [Am]. 2005;30:795-7.